There is a brand new dataset that Wormminator commited in one go last week and was kind enough to share.
Five instruments (yielding 10 pairwise combinations) measured each of the 35 flashlights.
-
On the plots below, each dot represents a different flashlight as measured by two different lightmeters.
-
Dot position on X-axis shows the mid-point of the two actual readings for each flashlight.
-
Dot position on Y-axis shows how far apart the two readings for the same flashlight were.
-
If any two instruments perfectly agreed, all dots would be on the gray dotted zero line.
-
If one instrument always read higher or lower than the other, but neither introduced any noise (the instruments perfectly agreeing with themselves), all the dots would be on the black solid bias line.
-
If one or both instruments were noisy, the dots will show as scattered clouds centered on gray zero line - but read the p.s. below.
-
If in addition to noise there is also a systematic bias between the two instruments, the cloud shape stays the same but the cloud’s centre will move up or down to the black bias line.
-
The plots allow to separate global bias from the inevitable scatter of paired disagreements and to estimate the contributions of each to the uncertainty of any future measurement.
-
If the overall systematic bias between pairs of known instruments can be estimated (like with this dataset) it can be corrected for, bringing the future readings of two instruments closer together.
-
Even if the global bias is estimated and corrected for, unremovable scatter will remain and the two future readings will still differ every time a pair of measurements is taken, but each time by a different amount and in a different direction.
-
If, for example, all I want is to compare tint readings by Opole4 vs. Seconic, I only need to look at plots 4a and 4b and ignore all other plots.
-
For high CRI lights Opple always reads higher overall - a few mduv higher for cooler lights and about twice as much for warmer lights. Even if I subtract this bias from all Opole readings from now on, to unbias them globally, the individual Opple reading will still commonly show tints up to ±2-5 mduv lower or higher than Seconic would, and unpredictably at that, even for identical lights.
-
For low CRI lights, the Opple’s readings are more consistently lower (more magenta) then what Seconic reports, by 2-3 mduv.
-
If we don’t know what instruments were used and take their reported readings at the face value (common case), both bias and scatter contribute to uncertainty about the difference between them (gray band).
A simple way of reading all these plots is to pretend there is no points and no x-axis and just look at the vertical size and location of the shaded zone on y-axis. The gray band shows the magnitudes of typical mismatch between the two instruments. Then look at points, lines, and patterns to get the rest of the story.
The values are sobering.
Note: I’m convinced that reporting, thinking about, comparing, and discussing fractions of mduv makes little sense given both, the uncertainties involved and the minimum difference that can be noticed.
The decimals above and in the plots’ legends are shown to reduce round off errors when adding or otherwise manipulating the values to get the answer.
Always round the final report(s) to the nearest integer, then check the plots to determine how many mduvs warrant treating the difference as a real signal rather than dismissing it outright as noise.
p.s. A word of caution: if you look at e.g. plots 8a or 8b, they look remarkably scatter-free. But the plots show the scatter of discrepancies between the two meters, not their individual noise.
It may be that both instruments are super smooth and noiseless and thus show nearly perfect agreement.
But it can also be that they are both properly noisy like all the other meters - they are just ‘noisy together’ and looking at the differences in their readings will not show scatter because the noise, while still random and unpredictable, is (inadvertently) synchronised between the two meters. And it’s not just for two units from the same maker - there may be other pairs like this.