Modifying Digital Multi Meter (DMM) Leads

What kind of switch is this?

maybe this

It is a solid switch, measured resistance only 5 mohm.
By the way as I see, the resistance of the tailcap depends mostly on the pressure spring.

Some resistance measurements, all done with a 4 terminal ohm meter (Resolution 0.01 mohm):

JetBeam M2S Tailcap 70 mohm
ThruNite TN11 Tailcap 85 mohm
4Sevens Quark X AA2 Tailcap 82 mohm
Fluke 289 10A range 33 mohm
Fluke 189 10A range 55 mohm
Fluke 179 10A range 35 mohm
Vichy VC99 20A range 48 mohm
Best DT9205A 20A range 18 mohm
Fluke test lead (x1) 38 mohm
Fluke test lead (x1) 51 mohm
No name test lead (x1) 257 mohm
No name 10A test lead (x1) 22 mohm

Both test leads and DMM's has connections included in the resistance (one for lead, two for DMM).

The Best DMM is unfused, that is the reason for the low resistance.

Remember when using test leads, that the resistance is included twice, one for each lead.


Doing a test with 10A current and measuring the voltage from test probe tip to test probe tip:

10A Tot. Res
Fluke 289 + Fluke test lead 1.02 volt 102 mohm
Fluke 189 + Fluke test lead 1.16 volt 116 mohm
Fluke 179 + Fluke test lead 0.97 volt 97 mohm
Best DT9205A + No name 10 test lead 0.59 volt 59 mohm

The connection resistance from test probe tip to current source is not included.

It seems I've been underestimating the tailcap resistance indeed.

And the above values don't include the thread connection, do they?

SC51 is IMO very prone to this.

You can actually notice from the lack of lumens when it does this.

I use Nyogel. Maybe I should run it dry from lubes.

My measurements are without any connection resistance. Two of the tailcaps has visible springs and I did connect as close to the tip as I could.

Tailcap to body resistance will be very low if the metal is clean. A LD01 SS body to head connections is below 1 mohm (This is a SS to SS connection, not aluminium).