The proper way of calculating lux / cd / candlepower

41 posts / 0 new
Last post

Pages

Enderman
Enderman's picture
Offline
Last seen: 1 day 18 hours ago
Joined: 11/03/2016 - 22:42
Posts: 3965
Location: Vancouver, Canada
The proper way of calculating lux / cd / candlepower

 

 

 

 

NOTICE

This method has been found to be incorrect, thanks to the tests of another forum member as well as my own. Please see these posts for more information:

http://budgetlightforum.com/comment/1053206#comment-1053206

http://budgetlightforum.com/comment/1281003#comment-1281003

NOTICE

 

 

 

 

 

 

 

 

 

 

 

 

 

So I found a pretty old post on the other flashlight forum about why just taking a random lux measurement and then calculating it back to 1m is wrong.

I did some research, and yes, it makes sense both mathematically and in practice.

 

Before doing all this math, I wanted to see for myself irl, so I took my tiny P5r.2 aspheric flashlight and put it horizontally.

I was very careful to get all the measurements right, and the peak lux in the hotspot:
At 1.00m I measured 12000 lux (12000 cd).
At 2.00m I measured 4500 lux. Now logically, based on 1/r^2, you can calculate this back to 1m by 4500*(2^2)/(1^2) and you get..........18000 LUX???

YES that is correct! The farther you measure lux from, the higher the cd value you get!
At 2m shouldn't it have been 3000 lux? Let's see why not:

 

 

So now, let me explain why almost every lux/cd/candlepower you have ever seen is (somewhat) wrong:

 

First of all, there is this scary-sounding thing called etendue: https://en.wikipedia.org/wiki/Etendue

It seems complicated, but the main thing you need to know is just the inverse law of light:

 

Pretty simple right? 2x distance, 1/4 lux.

Now, the reason for this is because light is emitted in a 360 degree sphere. Taking a look at a slice of a sphere, you see this:

 

Looking good?

Well here comes trippy part:

The head of a flashlight is NOT a point. The beams coming out of (most) flashlights is NOT spreading out by 1/r^2 (the spill is, but the beam is not)

 

Imagine a perfect thrower, perfectly collimated light ways, with 0 degrees divergence. The head of the light is 100mm, the spot at any distance is 100mm.

Ignoring the resistance of air in the way, the lux at 1m would be exactly the same as the lux at 10m, or 100m, because the spot is not increasing in size. All photons at 1m are the same as the number of photons at 100m.

--------------------------------------------------------------------------------------------------------------------------------------------------

Now imagine a flashlight like aspheric or large reflector throwers, which are CLOSE to collimated:

As you can see, the light coming out of a flashlight is actually a SECTION of the inverse square light law, a section that does not begin at r=0.

 

This means that the light is diverging LESS than if the flashlight was the source. How much less will depend on how well your flashlight collimates the light.

This is easy to calculate by just measuring how wide the beam is at the head of your flashlight, and how wide the spot is at a distance.

-----------------------------------------------------------------------------------------------------------------------------------------------------------------------------

Fortunately, there is another really smart person who made a web calculator for this: http://nightsword.com/uniformbeamcalc/

But at first I didn't trust it, so here is some math to check:

 

Say we have a flashlight with a 100mm lens, and the full lens is being used, so there is a 100mm diameter beam coming out of the flashlight.

The beam is fairly collimated by the lens, and I measure a 300mm spot at 10m distance, with 50000 lux.

The point of the triangle is behind the light by some distance we do not know, so we find it using some simple trig that you probably learnt about in high school:

Except we know the tree is 300mm tall and the person is 100mm tall, and the ditance between the tree and the person is 10m.

(you might be thinking I should have used 150mm and 50mm, but the truth is the ratio between 300 and 100 is the same as the ratio of 150 to 50, and only the ratio matters!)

so 300/100 = (10+x)/x where x is the distance behind the light

3x = 10+x

2x = 10

x = 5

The light rays converge to a point 5m behind the light, and the 10m you measured at is actually 15m from the origin.

NOW we take the 50000 lux (50k lux) we measured and apply the inverse law of light,

(lux @ 1m) / (50000) = (15m)^2 / (1m)^2

lux @ 1m = 225 / 1 * 50000

lux @ 1m = 11 250 000 lux = 11.25M lux = 11.25M cd = 11.25M candlepower

Since some american made that calculator, we are inconvenienced and need to use google to convert m to feet and mm to inches....

BUT IT WORKS!

 

----------------------------------------------------------------------------------------------------------------------------------------

Why you might not want to care:

-lights with a small head diameter have pretty large beam divergence, just look at any reflector EDC flashlight, which means the convergence point behind the flashlight is not that far behind, so lux measurements are fairly accurate.

-the farther the lux is measured at the more accurate it is (basic trig, imagine the triangle above, the ratio of measurement distance/total distance gets closer to 1), so it's not like you're actually getting less lux from your light, the rating on the light is an underestimate

-maybe you don't care about lux/cd throw at all Silly

--------------------------------------------------------------------------------------------------------------------------------------

Why I care:

-I have been underestimating my lux output of my aspheric throwers. I need to go outside and measure the spot size at about 50-100m distance, then properly calculate the candlepower/cd/lux.

-measuring indoors is not very accurate because of the way lenses work, you can end up with incorrect values due to the rays crossing over eachother if not focused correctly, which is why accurate results should use a measurement of "a few hundred feet" as it says on the website Smile

-I hope you learnt something, I certainly did, and I'm looking forward to finding out the real values of my flashlight.

Edited by: Enderman on 02/20/2018 - 21:40
sac02
sac02's picture
Offline
Last seen: 1 day 4 hours ago
Joined: 09/26/2016 - 15:18
Posts: 757
Location: TX

Makes a lot of sense. I had kind of wondered about that myself when measuring from 5 meters, a 4inch spot on the wall that came from a 2inch lens.

But I foresee difficulty changing the longtime “standard” way people measure/calculate lux/cd. Though the fact that the new “correct” method results in better numbers should help make people want to adopt it. Silly

Terry White
Terry White's picture
Offline
Last seen: 1 year 5 months ago
Joined: 09/22/2016 - 04:14
Posts: 1018
Location: South Carolina

Ummmm..... okay....

 

If it's broke I can fix it... From now on if  it's math I'm sending it to you...

Terry White

Skype terrywhite1@live.com

Koenta
Offline
Last seen: 1 year 3 months ago
Joined: 08/23/2016 - 08:16
Posts: 46
Location: Netherlands

Hello Endermann,

Very nice research and good explanation Smile

Terry White
Terry White's picture
Offline
Last seen: 1 year 5 months ago
Joined: 09/22/2016 - 04:14
Posts: 1018
Location: South Carolina

My math is not as bad as I make it sound. However I do use a phone app for the calculations. It's an app I use for work mostly that happens to include lighting calculations lol.

Terry White

Skype terrywhite1@live.com

Enderman
Enderman's picture
Offline
Last seen: 1 day 18 hours ago
Joined: 11/03/2016 - 22:42
Posts: 3965
Location: Vancouver, Canada

sac02 wrote:
Makes a lot of sense. I had kind of wondered about that myself when measuring from 5 meters, a 4inch spot on the wall that came from a 2inch lens.

But I foresee difficulty changing the longtime “standard” way people measure/calculate lux/cd. Though the fact that the new “correct” method results in better numbers should help make people want to adopt it. Silly


Most flashlights are not that highly collimated so the lux measurements are “accurate enough”, and you can’t get sued for it because the true numbers are actually higher, the numbers on the box are an underestimate.
If everyone agreed to measure lux at 10m or something standard like that then it would be easier to compare between flashlights, but that’s not gonna happen Silly

Terry White wrote:

Ummmm….. okay….


If it’s broke I can fix it… From now on if  it’s math I’m sending it to you…


Hahaha and I’m not even a math major xD
Thanks though

Koenta wrote:
Hello Endermann,

Very nice research and good explanation Smile


Thank you Smile
Terry White
Terry White's picture
Offline
Last seen: 1 year 5 months ago
Joined: 09/22/2016 - 04:14
Posts: 1018
Location: South Carolina

My mistake even using the app has always been either the room was not totally dark or I had my distance wrong. And the angle... well it's always wrong

 

 

Terry White

Skype terrywhite1@live.com

Enderman
Enderman's picture
Offline
Last seen: 1 day 18 hours ago
Joined: 11/03/2016 - 22:42
Posts: 3965
Location: Vancouver, Canada

Terry White wrote:

My mistake even using the app has always been either the room was not totally dark or I had my distance wrong. And the angle… well it’s always wrong


I don’t turn off the lights when taking my measurements, when I set the luxmeter range to x100 the tiny 50lux from my room lights become 0 and do not affect the 50000 lux from my light Silly
So for me it’s an insignificant decimal place that doesn’t even get considered.
JamesB
Offline
Last seen: 51 min 41 sec ago
Joined: 08/24/2011 - 14:43
Posts: 866
Location: France

I see your point bu i somewhat remember having the exact same number when measuring my B158 at 5 and 100 meters using the simplified widespread method of calculation, maybe i should try it another time.

I was under the impression that the common wisdom among us was that 1 or 2 meters readings may not be accurate and that around 5 meters was kind of the minimum distance to take readings.

EasyB
Offline
Last seen: 40 min 50 sec ago
Joined: 03/09/2016 - 15:24
Posts: 1710
Location: Ohio

I have to say that this possible effect has caused me some confusion regarding my understanding of throw and how to predict it, so I did some measurements to see what is actually happening.

My test light was a UF-1504, with 62mm diameter lens and XPL HI V2 1A. I put in a 8×7135 driver to keep the output more constant with time. I measured the lux at three different distances measured from the lens.

__d_______lux_____cd
6.05m____7230___264.6Kcd
8.34m____3840___267.1Kcd
10.29m___2530___267.9Kcd

These measurements are consistent with measuring the distance right from the lens and not at some point behind the lens.

Plugging the 10.29m measurement (the beam size at this distance was 36cm) into the calculator you linked (after converting units), results in:
divergence distance behind aperture: 2.14m
candlepower: 390.9Kcd

The calculator results are not consistent with my measurements and how they vary with distance. There are certainly some things in your explanation of the effect that make sense, but actual measurements tell a different story.

A different effect could have caused your conflicting results when you measured the lux at 1m and 2m. At small distances like 1m, the lens might not be filled with the image of the LED, from the point of view of the lux meter. Moving farther back to where the lens is filled would then result in a larger throw number.

Enderman
Enderman's picture
Offline
Last seen: 1 day 18 hours ago
Joined: 11/03/2016 - 22:42
Posts: 3965
Location: Vancouver, Canada

JamesB wrote:
I see your point bu i somewhat remember having the exact same number when measuring my B158 at 5 and 100 meters using the simplified widespread method of calculation, maybe i should try it another time.

I was under the impression that the common wisdom among us was that 1 or 2 meters readings may not be accurate and that around 5 meters was kind of the minimum distance to take readings.


I’m pretty sure the reason people don’t measure at 1m is because a luxmeter can only do like 20000 or 200000 lux, and at 1m an aspheric flashlight is above that by a lot.

EasyB wrote:
I have to say that this possible effect has caused me some confusion regarding my understanding of throw and how to predict it, so I did some measurements to see what is actually happening.

My test light was a UF-1504, with 62mm diameter lens and XPL HI V2 1A. I put in a 8×7135 driver to keep the output more constant with time. I measured the lux at three different distances measured from the lens.

__d_______lux_____cd
6.05m____7230___264.6Kcd
8.34m____3840___267.1Kcd
10.29m___2530___267.9Kcd

These measurements are consistent with measuring the distance right from the lens and not at some point behind the lens.

Plugging the 10.29m measurement (the beam size at this distance was 36cm) into the calculator you linked (after converting units), results in:
divergence distance behind aperture: 2.14m
candlepower: 390.9Kcd

The calculator results are not consistent with my measurements and how they vary with distance. There are certainly some things in your explanation of the effect that make sense, but actual measurements tell a different story.

A different effect could have caused your conflicting results when you measured the lux at 1m and 2m. At small distances like 1m, the lens might not be filled with the image of the LED, from the point of view of the lux meter. Moving farther back to where the lens is filled would then result in a larger throw number.

That’s interesting. Is the entire lens actually being used to focus the light?
I usually check by making sure there is led light spilling off the sides of my lens, and using a water sprayer to put water in the air so I can make sure the beam is filling the lens.
Your results still show that the farther you measure at the higher the cd you get. Maybe try measuring at 50 or 100m? Silly
I will need to do more tests with different lights to double check.

djozz
djozz's picture
Offline
Last seen: 1 hour 1 min ago
Joined: 09/07/2012 - 17:04
Posts: 16000
Location: Amsterdam

I really can not follow your story very well, I think that you are simplifying the direction that the rays coming out of a flashlight as if they start from a point source which certainly is not the case, flashlight optics and hotspot size/brightness have everything to do with the led not being a point source.

Hotspot brightness has to do with the fact that flashlight optics are imaging a led at a distance, so it more or less follows the lens maker’s formula creating an image of the led at a distance, smaller led or longer focal distance of the lens=smaller image, . Even reflectors do something similar although the way they ‘image’ is way more complicated.
One consequence is that if the image is created at twice the distance from the lens, the image is twice as large (4 times the surface area) and is thus 4 times less bright.

And my aspheric lights follow that simple: twice the distance from the lens= 1/4 the brightness pretty accurately, even when starting from 1 meter.

Reflector lights need more distance to form their uniform hotspot (they are not focusable like zoomies), but my usual 7 meters indoors is enough, and the simple relation between distance and brightness is followed.

Enderman
Enderman's picture
Offline
Last seen: 1 day 18 hours ago
Joined: 11/03/2016 - 22:42
Posts: 3965
Location: Vancouver, Canada

djozz wrote:
I really can not follow your story very well, I think that you are simplifying the direction that the rays coming out of a flashlight as if they start from a point source which certainly is not the case, flashlight optics and hotspot size/brightness have everything to do with the led not being a point source.

Hotspot brightness has to do with the fact that flashlight optics are imaging a led at a distance, so it more or less follows the lens maker’s formula creating an image of the led at a distance, smaller led or longer focal distance of the lens=smaller image, . Even reflectors do something similar although the way they ‘image’ is way more complicated.
One consequence is that if the image is created at twice the distance from the lens, the image is twice as large (4 times the surface area) and is thus 4 times less bright.

And my aspheric lights follow that simple: twice the distance from the lens= 1/4 the brightness pretty accurately, even when starting from 1 meter.

Reflector lights need more distance to form their uniform hotspot (they are not focusable like zoomies), but my usual 7 meters indoors is enough, and the simple relation between distance and brightness is followed.


But the lens is not producing a point source of light…
That’s why applying the inverse law of light assuming that the lens is at r=0 is incorrect.

If you had an ideal light source behind a lens, the lens would collimate the rays and the intensity would not change by distance, which would be infinite candlepower.

JamesB
Offline
Last seen: 51 min 41 sec ago
Joined: 08/24/2011 - 14:43
Posts: 866
Location: France
Enderman][quote=JamesB wrote:
I’m pretty sure the reason people don’t measure at 1m is because a luxmeter can only do like 20000 or 200000 lux, and at 1m an aspheric flashlight is above that by a lot.

No, ever since i started this hobby back when flashlight where much less bright people pretty much agreed on the wisdom that the closer the measuring distance, the less accurate, one or two meters readings have always been considered irrelevant and met with encouragment to measure from farther as far as i can remember, absolutly nothing to do with luxmeter capabilities.

EasyB
Offline
Last seen: 40 min 50 sec ago
Joined: 03/09/2016 - 15:24
Posts: 1710
Location: Ohio

Enderman wrote:

That’s interesting. Is the entire lens actually being used to focus the light?
I usually check by making sure there is led light spilling off the sides of my lens, and using a water sprayer to put water in the air so I can make sure the beam is filling the lens.
Your results still show that the farther you measure at the higher the cd you get. Maybe try measuring at 50 or 100m? Silly
I will need to do more tests with different lights to double check.

Yes, the entire lens is filled with the image of the LED for the distances I measured. I check by looking into the beam at different distances using sun viewing glasses.

EasyB
Offline
Last seen: 40 min 50 sec ago
Joined: 03/09/2016 - 15:24
Posts: 1710
Location: Ohio
Enderman wrote:
I’m pretty sure the reason people don’t measure at 1m is because a luxmeter can only do like 20000 or 200000 lux, and at 1m an aspheric flashlight is above that by a lot

I think it is because of the reason I mentioned above; at close distances the lens is not filled with the image of the LED. You cannot tell this by spraying water; you have to actually look into the beam at the lens.

DB Custom
DB Custom's picture
Offline
Last seen: 18 min 3 sec ago
Joined: 01/13/2013 - 22:28
Posts: 20324
Location: Heart of Texas

There are ANSI standards, developed for and by the lighting industry by specialists in the field. There is probably a reason they don’t use if’s and assuming in their definitions. The primary objective is to establish parameters of comparison, to create an equal and repeatable playing field for the purpose of recreative ability in determining the value of one light versus another as suited to a particular task. At the end of the day, deviation from the accepted standard is disqualified summarily as non applicable to the established criteria.

So, in simple terms, if we don’t all use the same test, our results cannot be compared and as such, the tests themselves are disqualified.

Don’t sell us, sell ANSI, then we will adapt accordingly.

An Enderman’s primary purpose is to create chaos and confusion.

Lexel
Lexel's picture
Offline
Last seen: 37 min 9 sec ago
Joined: 11/01/2016 - 08:00
Posts: 5489
Location: Germany

it is irrelevant where the imagine source of the flashlight is
you get only higher lumens if you go to a negative distance towards the spot where the light should for a focal point,
but that is physically impossible

if the light has 20mrad divergence it doesnt matter where you put a ideal luxmeter as long as the beam is uniform in brightness,
as a reflector light is not uniform in brightness at close distances, so you have to go further away,
if you are too close you measure too low lumens in the donut hole or too high lumens in the ring where the brightness is concentrated

on a reflector light the spill also affects the measured lux if you are very close

the next factor is that your measurement would need to be taken on an indefinate small point in the middle of the beam,
so the bigger the spot is the less you get a false reading as your luxmeter has a quite big spot to catch light at low distances

Enderman
Enderman's picture
Offline
Last seen: 1 day 18 hours ago
Joined: 11/03/2016 - 22:42
Posts: 3965
Location: Vancouver, Canada

Lexel wrote:
it is irrelevant where the imagine source of the flashlight is

It does though..
If you take a measurement at 10m you would think at 20m it would be 1/4 brightness, right?
But if the convergence point behind the flashlight is 100 000m behind, then the difference between 10m and 20m is nearly 0 because 100 010^2 and 100 020^2 are almost the same number.
Unlike 10^2 and 20^2 where the second number is 4x bigger.

Hopefully the 100000m exaggeration is enough for you to picture what is happening here.

Terry White
Terry White's picture
Offline
Last seen: 1 year 5 months ago
Joined: 09/22/2016 - 04:14
Posts: 1018
Location: South Carolina

Lol I need you to crunch some numbers for me. Then in about a week there will be a tiny upgrade to 1 of my sk68 and 1 sk98 lights. I'm curious what of any difference will be. 

 

I have some higher output  cells coming. I'm told I should notice a difference.

 

If you are up to it, my light meter is built into my phone app. Just let me know at what distance or whatever you may need.

 

Let me know if you are up to it. I will be typing up a review so it may take me a few minutes to respond. And if you don't feel up to it that's olay too. I know all about being tired. I'm very good at it.

Terry White

Skype terrywhite1@live.com

Enderman
Enderman's picture
Offline
Last seen: 1 day 18 hours ago
Joined: 11/03/2016 - 22:42
Posts: 3965
Location: Vancouver, Canada

Terry White wrote:

Lol I need you to crunch some numbers for me. Then in about a week there will be a tiny upgrade to 1 of my sk68 and 1 sk98 lights. I’m curious what of any difference will be. 


 


I have some higher output  cells coming. I’m told I should notice a difference.


 


If you are up to it, my light meter is built into my phone app. Just let me know at what distance or whatever you may need.


 


Let me know if you are up to it. I will be typing up a review so it may take me a few minutes to respond. And if you don’t feel up to it that’s olay too. I know all about being tired. I’m very good at it.


The more the better, because the margin of error when calculating lux and measuring the spot size is lower.
Although it should be in the tens of meters, not hundreds, because then the lux values are too low to be read by the meter Silly
Lexel
Lexel's picture
Offline
Last seen: 37 min 9 sec ago
Joined: 11/01/2016 - 08:00
Posts: 5489
Location: Germany

Enderman wrote:
Lexel wrote:
it is irrelevant where the imagine source of the flashlight is

It does though..
If you take a measurement at 10m you would think at 20m it would be 1/4 brightness, right?
But if the convergence point behind the flashlight is 100 000m behind, then the difference between 10m and 20m is nearly 0 because 100 010^2 and 100 020^2 are almost the same number.
Unlike 10^2 and 20^2 where the second number is 4x bigger.

Hopefully the 100000m exaggeration is enough for you to picture what is happening here.

you are wrong
if the convergence point would be 100000m behind the flashlight just the divergence would be really low about 0.000001mrad if the lens diameter is 10cm
at 0.000001mrad the Lux value between 10 and 20m would be almost the same
lets say at 1m I got 100cd
now I got to to 10km

you are just not understanding the difference between a flashlight and a laser

take a laser pointer with 3mm beam and 0.25mrad single Mode diode

I measure at 0m 10000cd

at 1m 3mm+(0,25mrad*1m)=3,25mm spot
calculated 8521cd

at 2m 3mm+/0,25mradx2m)=3,50mm spot
calculated 7346cd

I got 15 years experience with lasers also calculating things trust me

scianiac
scianiac's picture
Offline
Last seen: 5 months 1 week ago
Joined: 02/05/2015 - 18:50
Posts: 211
Location: Northeast, US

I understand the principle here but I can see why it’s sorta hard to explain and actually using it is also quite a trick. I have a question for EasyB though that may clear some things up, how did you measure the spot, it is of course square so did you measure it’s width or diagonal. I think the correct measurement should be diagonal if you are using the lens diameter as the aperture.

I’m going to do some more extensive tests shortly but as I’ve thought about this in the past I sorta came to the same conclusion as DBC in that in the end it doesn’t matter because numbers are great but if everyone else is just measuring lux at at distance then you are “cheating” kinda, I mean you may be more accurate but for instance in my case selling a light using this more “correct” math means that when people measure it they will come up short and it will be hard to compare with other lights measured in the same fashion. As in for personal uses it’s fine but it’s hard to move out from there.

Then there is the measurement problem, it’s not exactly easy to get really accurate hotspot sizes, sure you’ll be close but there is a lot of messing about and guessing required since the beam is not clearly enough defined at that range. You would have to move the lux meter around to chart the lux change.

However, because this effect gets more and more pronounced the higher the kcd using the “basic” measuring setup may cause some issues because the “basic” measurements will rapidly start generating higher kcd with longer and longer measurement distances. So maybe this is really needed for 1Mcd+ lights. I’m going to try to get some different tests done with our 1-3Mcd prototypes and see if I can get some good comparison numbers.

Co-owner/Engineer at STO Flashlights.

Enderman
Enderman's picture
Offline
Last seen: 1 day 18 hours ago
Joined: 11/03/2016 - 22:42
Posts: 3965
Location: Vancouver, Canada
Lexel wrote:
the divergence would be really low about 0.001mrad if the lens diameter is 1cm at 0.001mrad the Lux value between 10 and 20m would be almost the same

EXACTLY
THIS IS MY POINT Smile

You can’t take a lux measurement at 20m and then just calculate it back to 10m by 1/r^2 by doing lux*20^2/10^2
Because the flashlight is not where r=0 is!
If the divergence is really low, just as you said, the lux will barely be any different! It will not be 4x higher at 10m compared to 20m!

Enderman
Enderman's picture
Offline
Last seen: 1 day 18 hours ago
Joined: 11/03/2016 - 22:42
Posts: 3965
Location: Vancouver, Canada

scianiac wrote:
I understand the principle here but I can see why it’s sorta hard to explain and actually using it is also quite a trick. I have a question for EasyB though that may clear some things up, how did you measure the spot, it is of course square so did you measure it’s width or diagonal. I think the correct measurement should be diagonal if you are using the lens diameter as the aperture.

I’m going to do some more extensive tests shortly but as I’ve thought about this in the past I sorta came to the same conclusion as DBC in that in the end it doesn’t matter because numbers are great but if everyone else is just measuring lux at at distance then you are “cheating” kinda, I mean you may be more accurate but for instance in my case selling a light using this more “correct” math means that when people measure it they will come up short and it will be hard to compare with other lights measured in the same fashion. As in for personal uses it’s fine but it’s hard to move out from there.

Then there is the measurement problem, it’s not exactly easy to get really accurate hotspot sizes, sure you’ll be close but there is a lot of messing about and guessing required since the beam is not clearly enough defined at that range. You would have to move the lux meter around to chart the lux change.

However, because this effect gets more and more pronounced the higher the kcd using the “basic” measuring setup may cause some issues because the “basic” measurements will rapidly start generating higher kcd with longer and longer measurement distances. So maybe this is really needed for 1Mcd+ lights. I’m going to try to get some different tests done with our 1-3Mcd prototypes and see if I can get some good comparison numbers.

Good point!
The farther you calculate it, the closer you get to the true value, but users usually do less than 20m measurements so they would get a smaller number than yours.

Measuring the spot size is hard unless you have a highly collimated light with a sharp projection, so tbh this method is not really useful for typical flashlights and reflector lights, it is more for aspheric throwers and sky beams and stuff like the maxabeam.

Also, IIRC the maxabeam was measured at 3M lux at 1meter, but the candlepower is 12 million. So yeah, goes to show how misleading lux can be when calculated incorrectly for those types of lights.

Lexel
Lexel's picture
Offline
Last seen: 37 min 9 sec ago
Joined: 11/01/2016 - 08:00
Posts: 5489
Location: Germany

if the convergence point would be 100000m behind the flashlight just the divergence would be really low about 0.000001mrad if the lens diameter is 10cm
lets say at 1m I got 100cd

now I got to to 100.000m then my spot is exactly 20cm
so at 100km of the flashlight I get 25cd
if I calculate that lux to 1m I get 25,000,000,000cd
if I measure at 300.000km or one light second
I get a spot of 600m
0,0000027cd
If I calculate that to 1m
I get 250,000,000,000cd

so for this almost perfect thrower the distance to get an accurate reading would be a couple of light seconds away

there will be a point where the cd doesnt raise anymore
how far that is away depends on the divergence and lens diameter

as flashlights are really bad on divergence that point is really close to the light 5-10m should be fine

just go further out and when the calculated CD value dont increese anymore the reading is accurate

you are just not understanding the difference between a flashlight and a laser

take a laser pointer with 3mm beam and 0.25mrad single Mode diode

I measure at 0m 10000cd

at 1m 3mm+(0,25mrad*1m)=3,25mm spot
measured 8521cd

at 2m 3mm+/0,25mradx2m)=3,50mm spot
measured 7346cd
if I calculate that to 1m I would get 29384cd

at 100m 3mm+(0,25mrad*100m)=28mm
measured 115cd
if I calculate that to 1m I would get 1,150,000cd

at 100km 3mm+(0,25mrad*100,000m)=25003mm
measured 0,00144cd
if I calculate that to 1m I would get 1,440,000cd

for a Laser that will be in the range of a couple hundred meters to kilometers
it all depends on ratio for divergence and lens diameter

I got 15 years experience with lasers also calculating things trust me

Enderman
Enderman's picture
Offline
Last seen: 1 day 18 hours ago
Joined: 11/03/2016 - 22:42
Posts: 3965
Location: Vancouver, Canada

Lexel wrote:

as flashlights are really bad on divergence that point is really close to the light 5-10m should be fine

just go further out and when the calculated CD value dont increese anymore the reading is accurate

you are just not understanding the difference between a flashlight and a laser


Yes good job, now you’re basically saying the same thing I did in the original post.
For most people with regular flashlights, it doesn’t really matter.

For HIGHLY COLLIMATED lights such as aspheric lights or short arc lights (eg the maxabeam) then YES there is a significant difference between just measuring lux at a few hundred feet vs actually calculating the real candlepower due to the highly collimated sub-1 degree of divergence.

It does NOT need to be a laser, anything that partially collimates light is affected by this property.
This is why if I measure my light at 5m it gets a different cd rating than if I measure it at 10m or 15m.

I honestly don’t understand why you’re even trying to argue, you’re basically proving my point a second time…

scianiac
scianiac's picture
Offline
Last seen: 5 months 1 week ago
Joined: 02/05/2015 - 18:50
Posts: 211
Location: Northeast, US

I mean it’s not like there is some magic point where etendue compensation is suddenly needed, in effect you are both right the question is what magnitude are we talking here. Lexel is probably right up to surprisingly high kcd/lens size/collimation. Surely well in the range of even the best off the shelf “throwers”, or least that’s how it seems to me, seems like the error with them would be well within the numerous other factors involved. But after 1Mcd things start to get different, we aren’t dealing with the 3mm aperture Lexel used in his example, we are dealing 100mm+ lenses. The divergence is surprisingly small. I of course don’t have measurements yet but things sure look awfully small. If I recall that tree line is 1300m away. So I guess my seat of the pants opinion is that there is some compensation needed at these ranges and perhaps it’s enough to be worth taking into account.

Co-owner/Engineer at STO Flashlights.

Terry White
Terry White's picture
Offline
Last seen: 1 year 5 months ago
Joined: 09/22/2016 - 04:14
Posts: 1018
Location: South Carolina

Okay I just took a screen shot of the reading.

This is at 5 meters. Cree XML2  and I did not install this one. This is one I believe to actually be a real  Ultrafire, so no clue of bin or color

 I can tell you  it is the whitest of any XML2 I jave seen. Let me know what else you need . 

 

Terry White

Skype terrywhite1@live.com

Lexel
Lexel's picture
Offline
Last seen: 37 min 9 sec ago
Joined: 11/01/2016 - 08:00
Posts: 5489
Location: Germany

Normal flashlights have a hotspot of 10-20 degree with 1” diameter

The virtual sorce point is often not even behind the body of the light, if its a short reflector
We are talking here about 10-30cm, if you measure at 3-5m the difference is very small

A extreeme thrower has less divergence and big head
Both things push it towards bad readings at low distances
Lets say you get 2 degree with 10cm head the virtual point is less than 3 meters behind the flashlight

The fact that your luxmeter has a big aperatur catching light compensates the virtual source behind the front of the flashlight, ideally the luxmeters aperature should be 0mm2
So often the measurement at too low distance is still correct as both values negate each other

The whole calculating cd to 1m is not really useful to get the real brightness if there is not a defined distance for measurement set for beam angle/head diameter, as well how big the aperature of the luxmeter has to be

EasyB
Offline
Last seen: 40 min 50 sec ago
Joined: 03/09/2016 - 15:24
Posts: 1710
Location: Ohio

scianiac wrote:
I have a question for EasyB though that may clear some things up, how did you measure the spot, it is of course square so did you measure it’s width or diagonal. I think the correct measurement should be diagonal if you are using the lens diameter as the aperture.


Yes I used the width. Not sure whether the width or diagonal is the correct measurement, but it doesn’t make a big difference. My measurements are still inconsistent with the model proposed here.

I’m not convinced the model is accurate for our flashlights, at least not to the extent to which that calculator is predicting. We should be able to test the predictions, and, again, my measurements show them to not be accurate.

We need some more carefully done measurements on largish diameter lights. The tester should be careful to keep the output the same for each measurement. Also there is another effect that I have mentioned that would change the measured cd; at distances too close to the lens, the lens is not filled by the image of the LED. The tester should only measure at distances where the entire lens is filled with light. (tested by looking into the beam at the lens from the distance in question) This distance for my 1504 lens with XPL HI was roughly 2m.