DMM Mastech MS8216
This is a small pocket DMM with voltage, frequency, ohms and capacity.
The meter was in a clamshell pack. On the back is a the specifications.
In the pack was the meter and a manual in English.
The probes are small and they are not round, but ellipse shaped.
The probes are directly connected to the meter, no plugs and sockets.
The range switch is easy to turn, but the detents are not very strong and it is very easy to place the switch between ranges.
A pouch is included.
Two screws must be removed to get battery access.
Display
The above picture shows all the segments on the display.
Typical display during usage, it will show the selected range and value.
Functions
Buttons:
Rel: Store current display value and show all new values relative to this value.
H/D: This is short for Hz/duty cycle and will show frequency and duty cycle in both DC, AC and frequency mode.
D-H: This is the hold button and will freeze the display reading, until pressed again.
Selection of Rel and Hz/duty will disable auto ranging and it stays disabled
Rotary switch:
Off: Meter is turned off
VAC: Voltage AC, frequency and duty cycle.
VDC: Voltage DC, frequency and duty cycle.
: Ohm. -
: Continuity. -
: Diode. -
: Capacitance. -
Hz: Logical frequency input.
Input
This meter only have a red and black probe comming out, no other connections.
Measurements
Volt and frequency
At 1V rms input on frequency the counter range is from 2Hz to 50kHz
At 2V rms input on frequency the counter range is from 0.6Hz to 100kHz
At 5V rms input on frequency the counter range is up to 290kHz
Duty cycle works from 15% to 90% at 1kHz with 4Vpp, precision is within 3.0
Frequency inputs do not need a zero crossing.
1V AC readings is 5% down at 2kHz
Voltage input impedance is 10-11Mohm,
mV input impedance is high below 1.5V and drops to 10Mohm above
Frequency input is similar to mV input.
Input protection is 600V AC/DC for voltage 250V AC/DC for frequency.
Current
Meter cannot measure current
Ohm, continuity, diode and capacity
Ohm needs about 3.4s to measure 100ohm
Ohm is 0.42V open and 0.17mA shorted
Continuity is moderate in speed (100ms)
Continuity beeps when resistance is below 55ohm
The beeper volume is very low.
Continuity is 0.44V open and 0.17mA shorted
Diode range uses 1.5V, max. display is 0.999V at 0.2mA, max. current is 0.55mA shorted
190uF takes about 40 seconds to measure.
Rated overload protection is 250 VDC/VAC
Miscellaneous
Current consumption of meter is 0.8mA except AC where it is 1.4mA
Meter fails below 1.4V, battery symbol show at 2.4V.
Readings will drift about 2 count when from when battery symbols shows to the meter fails.
Viewing angle is good, except from top
Display updates around 2 times/sec
Will automatic turn power off after about 15 minutes
The meter often need many display update to reach the final value.
Weight is 82g without accessories, but batteries.
Size is 110 x 78 x 13mm.
Probes
Probe wire is a bit thin, they are 31cm long.
Capacity measurement waveform.
Frequency input resistance in voltage frequency position, V and mVDC is similar.
Frequency input resistance in voltage AC position.
Lowest capacity range is sometimes out of tolerance (Up to 8%) and next time I measure the same capacitor it is inside tolerance.
Tear down
I had to remove two screws, to open it, this is also required when replacing batteries.
The battery holder is for two button cells.
I had to remove 6 screws before I could get the circuit board out.
The buzzer is mounted on the frontplate.
On this side there is only two capacitors and holes for adjusting the 3 trimpot (VR1, VR2, VR4).
On the input is a PTC (R24), it is not used for all ranges. The main input resistor (R21 & R’21 2x5Mohm) goes directly to the multimeter chip. My guess is that the input protection is a single transistor (Q1), it looks like the 5 diodes (D3..D7) is used for coding range switch to multimeter chip and the two other diodes (D1 & D2) may be rectifier circuit for AC measurement where VR2 is used to adjust it with.
The multimeter chip (IC1 CS7721) handles everything in the meter.
Conclusion
The meter is specified at 600V, but the ranges can only handle 250V in overload.
The meter has about the expected functions for a pocket meter without any current ranges. There is a few problematic areas with it: Range switch must be carefully placed in the correct position it do not snap into position, buzzer is difficult to hear, duty cycle is low performance and the lowest capacity range is not completely reliable.
Notes
How do I review a DMM